Posts: 15
Joined: Mon May 24, 2021 10:17 am

How To yose Mind monitor

Post by hiroko »

Normally, EEG voltage is said to be several μV to several tens of μV.
The data of RAW_TP9, RAW_AF7, RAW_AF8, and RAW_TP10 extracted from Mindmonitor sometimes show 700μV.

Are the RAW data (RAW_AF7, RAW_AF8, RAW_TP10) different from the EEG voltage?
User avatar
James
Site Admin
Posts: 1103
Joined: Wed Jan 02, 2013 9:06 pm

Re: How To yose Mind monitor

Post by James »

The RAW data is the EEG data for each channel on the headband as measured in microvolts.
Different manufacturers may normalize their results to different ranges, but Interaxon's is ~700uV as you state and this is correct.
In the FFT it's the variance that you're looking at. The mean value isn't going to affect that.